做13C为什么要用氘代试剂?

做13C为什么要用氘代试剂?试剂应该是12C取代的,当然,12C的天然丰度已经是98.9%,嘿嘿!
顶部
fence-straddler[使用道具]
五级
Rank: 5Rank: 5


UID 26191
精华 0
积分 1875
帖子 620
信誉分 101
可用分 2346
专家分 407
阅读权限 255
注册 2009-8-21
状态 离线
2
 

回复 #1 猴子 的帖子

需要氘代溶剂来完成锁场(Field lock)和匀场(deuterium gradient shimming).

In order to produce a high resolution NMR spectrum of a sample, especially one which requires signal averaging or phase cycling, you need to have a temporally constant and spatially homogeneous magnetic field. Consistency of the Bo field over time will be discussed here; homogeneity will be discussed in the next section of this chapter. The field strength might vary over time due to aging of the magnet, movement of metal objects near the magnet, and temperature fluctuations. Here is an example of a one line NMR spectrum of cyclohexane recorded while the Bo magnetic field was drifting a very significant amount.  The field lock can compensate for these variations.


The field lock is a separate NMR spectrometer within your spectrometer. This spectrometer is typically tuned to the deuterium NMR resonance frequency. It constantly monitors the resonance frequency of the deuterium signal and makes minor changes in the Bo magnetic field to keep the resonance frequency constant. The deuterium signal comes from the deuterium solvent used to prepare the sample. The animation window  contains plots of the deuterium resonance lock frequency, the small additional magnetic field used to correct the lock frequency, and the resultant Bo field as a function of time while the magnetic field is drifting. The lock frequency plot displays the frequency without correction. In reality, this frequency would be kept constant by the application of the lock field which offsets the drift.




On most NMR spectrometers the deuterium lock serves a second function. It provides the =0 reference. The resonance frequency of the deuterium signal in many lock solvents is well known. Therefore the difference in resonance frequency of the lock solvent and TMS is also known. As a consequence, TMS does not need to be added to the sample to set =0; the spectrometer can use the lock frequency to calculate =0.