Channel 3 : N15.
Channel 4 : H2 (Although we don't have the deuterium channel yet.)
On our unity 600 (victor), 4th channel was used or carbon before. In
this case, channels 1,2,3 and 4 correspond to 1H, 13C, 15N, and 13C. On
the other hand, in our unity+ 500MHz (miryl), channel 1,2,3, and 4 are
often assigned to H1, N15, C13, and C13. Therefore, there are three
types of experiments using 4 channels: the one for (1) victor, (2) miryl,
and (3) with 2H. We don't use (1) and (2) anymore.
Our Unity+ 500 MHz (miryl) share channel 4's hardware with that of
channel 3. You can not use channels 3 and 4 for different nuclei. In
Lewis's sequence with 2H option (there is an extention _D_ in the name
of the pulse sequences), the third channel is assigned to N15 while the
fourth is 2H. Therefore, set dn3='' (channel 4 equals non) and set
ampmode='dddd' when you use such pulse sequence currently. Ampmode is
sometimes defined to 'dddp' in the Lewis's parameter file. 'd' means
default mode while 'p' means pulse mode.
(8) How to get pulse sequences.
I strongly recommend you to get pulse sequences for 600MHz and for 500MHz each
from Lewis's group together with parameters. This is essential not to miss
experiments for you since parameters are different between 500MHz and 600MHz.
The pulse sequences available listed on cuturl('http://abragam.med.utoronto.ca.')
Send e-mail to
ranjith@bloch.med.utoronto.ca (Ranjith Muhandiram).
(9) Difference of shaped pulses created by Lewis and Pbox.
Lewis's shaped pulse is created by executing C programs installed in the
shapelib. Varian's original shaped pulses are in the ./nmr/wavelib and can
be modified further using Pbox. Follows is a comparison of near-identical
seduce pulses for carbon created by both methods.
1) Inversion seduce. These two creates the seduce pulse using almost
the same parameters as each other.
Pbox: { seduce 2000/0.000240 }
pw=240 us
power= 4.53 kHz
Lewis: RF_seduce 240 0 240 0 1023 0
pw=240
power= 4.53 kHz (by pulse tool)
2) seduce decoupling. The values of paramters used to create the pulse
are different between Pbox and Lewis. Especially, dmf in the Pbox
looks to be smaller than Lewis's one because of difference of dres.
Pbox: " { SEDUCE1 2000.0/0.0002400 }"
power= ~ 47 (power of 54 us as a hard 90 pulse)
B1 = 4.59
dmf=20833
dres =9
dmf*90/9=20833*90/9=208330
"DEC_seduce_cos 0 240"
-> pw_shape = 240 us
-> dres = 2
-> power = ~39 + 6 in our machine.
-> dres = 2
1/pw_shape=4166
dmf*90/dres=4166*90/2=187470.
NOTE: what's dres?
example:
1. dres=9, and dmf=40000
2. dres=90, and dmf=400000
1. and 2. are the same as each other. If you set dres = 90 then
dmf = 1/pw90, which makes it 1/stepsize. When dres is not 90, dmf is
not just 1/pw90. For example, when dres = 9 and dmf=1/pw90,
1/stepsize = dmf*90/9. See the manual for details.
** In the following explanation of individual experiments, I may not
particularly describe dres, dres2, and dressed. Typical values for dres
used in the Lewis's sequence are written in the nmrexp_hnco.txt.